Smart Pressure Transmitter Sensor Trim Guide with Diagrams & Calibration Steps

In industrial automation, smart pressure transmitters are key components for accurate and stable measurement of process pressure. They convert physical pressure into an electrical signal and feed it into the control system for monitoring and regulation. However, like any precision instrument, smart transmitters require calibration to maintain accuracy, especially after installation or over time.

One of the most effective digital calibration techniques used for these transmitters is called Sensor Trim. This process, along with Zero Trim and Analog Output Trim, ensures that the transmitter’s output is consistent with actual pressure conditions.

This guide will walk you through:

  • How sensor trim works
  • What zero and analog output trim mean
  • When and how to use low and high trims
  • Step-by-step interpretation of trim curves
  • Explanation of internal signal flow using custom diagrams

Let’s begin by understanding the internal working of a smart pressure transmitter.

Smart Pressure Transmitter Sensor Trim Guide with Diagrams & Calibration Steps - Smart Pressure Transmitter Signal Flow Diagram

The above diagram illustrates the internal signal path of a smart pressure transmitter from input to output. Here’s how it works:

  1. Pressure Applied: Process pressure is applied to the sensor diaphragm, which converts it into an analog electrical signal.
  2. ADC (Analog-to-Digital Conversion): The analog signal is converted into digital format by an Analog-to-Digital Converter (ADC).
  3. Microprocessor: This is the brain of the transmitter. It processes the signal, applies trim adjustments, damping, and range scaling (LRV, URV).
  4. DAC (Digital-to-Analog Conversion): After processing, the microprocessor sends a digital signal to the DAC, which converts it back into a 4-20 mA analog signal suitable for PLC or DCS systems.
  5. Trim Adjustments: You can apply both low and high trim at the sensor input side and also on the analog output to calibrate the entire path for maximum accuracy.

Refer to the first diagram again:

  • Sensor – ADC – Microprocessor – DAC – 4 – 20 mA output
  • Low/High Sensor Trim = Applied between sensor and ADC (corrects internal digital pressure)
  • Range Settings (LRV/URV) = Set at microprocessor level (maps pressure range)
  • Analog Output Trim = Applied after DAC (calibrates current loop)

This layered architecture is why modern smart transmitters require digital and analog calibration methods.

Sensor Trim is a digital calibration process used to align the transmitter’s internal sensor reading with a certified external pressure reference. It affects the internal characterization curve, which defines how input pressure relates to the process variable (PV) output.

Sensor trim doesn’t modify the analog output directly it corrects the raw sensor signal interpretation by the transmitter. It is vital when the transmitter starts to show drift or when replacing sensors/electronics.

Zero trim is a single-point calibration method that sets the transmitter’s reading to zero when exposed to atmospheric or equalized pressure. It’s mostly used to compensate for:

Live Zero vs Dead Zero: Understanding the Difference Between Live Zero and Dead Zero in 4 to 20 mA Signals

  • After installing a transmitter on a vertical or inclined pipe
  • After wet leg leveling or impulse line filling.
  • Before performing sensor trims.
  • Equalize the pressure by opening the equalizing valve.
  • Ensure both wet and dry legs are properly filled and leveled.
  • Apply zero trim via HART communicator or configuration software.

Best Practice: Perform zero trim before full sensor trim, only when no pressure is applied. Do not use zero trim across the full span. It should only be applied at the zero reference point to avoid altering the span of measurement.

Low trim adjusts the transmitter’s digital response at the lower range value (LRV) of the span. It’s always the first step in sensor trim calibration, and it shifts the entire curve up or down, correcting any offset.

Smart Pressure Transmitter Sensor Trim Guide with Diagrams & Calibration Steps - Low Sensor Trim - Correcting Offset Errors

In the above graph:

  • The orange line is the factory characterization curve showing inaccurate PV output.
  • The green dashed line is the corrected curve after applying the low trim.
  • The transmitter originally reads -152 inH₂O when -150 inH₂O is applied.
  • After applying the low trim, the entire curve is shifted parallel to match the correct value.

This correction improves accuracy across the span without changing the slope (span).

Master the Process: 6 Steps Guide to Selecting a Master Instrument for Calibration

High trim adjusts the transmitter’s response at the upper range value (URV). It modifies the slope of the characterization curve and compensates for gain errors.

Smart Pressure Transmitter Sensor Trim Guide with Diagrams & Calibration Steps - High Sensor Trim - Adjusting Span or Slope

In this graph:

  • After low trim has been applied, a known pressure of 150 inH₂O is applied.
  • The transmitter shows 152 inH₂O, indicating a gain or slope error.
  • Applying a high trim brings the output back to 150 inH₂O, and the new slope is shown by the green dashed line.
  • The span of the curve (difference between low and high) is corrected.

The combined effect of low plus high trim ensures full accuracy over the entire measuring range.

Free Access!: Free Instruments Calibration Procedures: 60+ Step-by-Step Methods for Pressure, Temperature, Flow & Level

After adjusting the internal sensor readings through sensor trims, the next step is to verify and calibrate the analog output. This is done via Analog Output Trim.

This process aligns the actual loop current output (4-20 mA) with the transmitter’s internal digital output. It ensures that the control system receives the correct signal even if there are slight hardware mismatches in the DAC or output circuitry.

Tools Required:

  • Accurate current meter (e.g., Fluke 744, 754)
  • Loop calibrator
  • HART communicator or manufacturer software

Use Case: When 4 mA doesn’t match LRV or 20 mA doesn’t correspond to URV even after sensor trims.

Don’t Overlook This!: Why Calibrating your Calibrators is Critically Important: Accuracy, Compliance and ISO 17025 and NIST Traceability

StepCalibration TypePurpose
1Zero TrimSet baseline with zero pressure
2Low Sensor TrimAlign lower end of sensor curve
3High Sensor TrimCorrect span/slope of sensor curve
4Analog Output TrimEnsure 4-20 mA matches digital PV
5ValidationCompare against reference & record results

Understand This First: Differences Between Validation and Calibration

FeatureSensor TrimAnalog Output Trim
AffectsDigital process variable4-20 mA current output
Internal AdjustmentYes (sensor-to-microprocessor)Yes (microprocessor-to-output)
PurposeMeasurement accuracySignal integrity
Required ToolsPressure calibrator, HART deviceCurrent meter, loop calibrator
When to PerformDuring pressure sensor calibrationDuring analog signal verification
  • Oil & Gas: Accurate pressure readings in custody transfer, wellhead monitoring, and flare gas systems are crucial. Sensor trims ensure deviations are eliminated due to sensor drift or harsh environments.
  • Power Plants: In boiler drum level or main steam pressure measurement, slight sensor inaccuracies can lead to major process inefficiencies. Sensor trim maintains measurement precision during high temperature cycles.
  • Pharmaceuticals: Sensor trim supports accurate CIP/SIP cycles by maintaining tight pressure tolerances in sterile systems. This ensures product safety and regulatory compliance.
  • Food & Beverage: For pasteurization and filling operations, consistent pressure readings ensure flow rates and batching stay within control limits, enabled by regular sensor and analog trims.

Step-by-Step Guide: Step-by-Step Procedure to Calibrate an Absolute Pressure Transmitter

In older systems, calibration only adjusted the analog output signal. However, smart transmitters require digital calibration (sensor trim) to maintain internal signal integrity. Modern devices feature onboard microprocessors and digital signal processing, so:

  • Sensor trim modifies how the transmitter interprets the pressure input
  • Analog trim adjusts how the signal is transmitted to control systems

Both are essential in a hybrid analog-digital environment to maintain reliable operations

ProblemLikely CauseSuggested Fix
Incorrect output after trimTrim sequence errorRepeat low to high to analog output trim
Current not matching displayLoop resistance or output errorRecheck analog output trim, wiring
High deviation after zero trimWet legs not filled or pressure not zeroEqualize pressure and retry
Pressure reading drifts over timeSensor drift or ambient temp effectsSchedule periodic trim
IssueCauseAction
Transmitter shows offsetZero trim not performedEqualize and perform zero + low trim
Reading accurate digitally, wrong outputAnalog signal mismatchPerform analog output trim
Curve slope appears distortedHigh trim skippedRepeat trim: low  then high then analog
Fluctuating outputPressure not stable or unfilteredApply damping, wait for stabilization
  • Always perform low trim before high trim
  • Use high-accuracy reference standards (NIST-traceable)
  • Stabilize pressure input before applying trim commands
  • Let the transmitter reach ambient temperature equilibrium
  • Use manufacturer-certified software (e.g., AMS, FieldMate, PACTware)
  • Document all “As Found” and “As Left” values for traceability

Top Picks: Best Calibration Management Software

  • Performing zero trim on a non-zero pressure input
  • Skipping low trim and only applying high trim
  • Using non-certified or inaccurate reference pressure sources
  • Neglecting analog output trim in hybrid control environments
  • Ignoring drift after environmental or temperature changes

Avoid These!: Top 15 Common Calibration Mistakes in Industrial Instruments

Sensor trim is a digital calibration technique that changes how the transmitter reacts to a known reference pressure. It precisely adjusts the output of the Analog-to-Digital Converter (ADC) to make up for any mistakes in the pressure sensor. This makes sure that the digital process variable (PV) appropriately shows the pressure that is being applied, which makes the measurements more accurate overall.

In the context of calibration, trimming means making small changes to a transmitter’s readings so that they match a certified reference standard. This could include:

  • Setting the lower and upper limits of the measurement range
  • Changing the sensor output (Sensor Trim)
  • Checking or fixing the analog output (4-20 mA Trim) Trimming makes sure that the sensor or transducer gives correct values all the way across its calibrated range.

Know the Difference: Why Calibration Isn’t the Same as Re-ranging in Process Instrumentation

A full sensor trim in a pressure transmitter is a two-point calibration that makes sure the sensor’s measurement curve is straight at both the low end and the high end.

The low trim is always done initially to get rid of the offset.

The high trim changes the measurement’s slope or span.

For instance, Yokogawa transmitters provide sensor trim using two main methods that make sure the performance stays the same across the pressure range.

Zero trim is a way to calibrate a transmitter so that it gives the right output when there is no pressure (a zero pressure reference). It moves the whole sensor characterisation curve up or down without changing its slope. This fixes offsets or mechanical imbalances that happen during installation.

  • Sensor Trim: Changes the internal reading by fixing the ADC output (the signal from the sensor to the CPU). It makes sure that the digital pressure reading is correct.
  • Output Trim (4-20 mA Trim): Changes the analog output signal so that the 4 mA (LRV) and 20 mA (URV) outputs match the internal digital reading. At the Digital-to-Analog Converter (DAC) stage, this is done.

Both types of trim are necessary in digital transmitters to keep the signal clean from sensor input to analog output.

Read More