Instrument Calibration Verification Calculator

An instrument calibration verification calculator confirms that instruments provide accurate and precise measurements by comparing their performance to a predefined standard. 

The following is a thorough example of a pressure transmitter calibration with five points.

Calibration Verification Calculator 1
  1. Instrument Reading: The value provided by the instrument being tested.
  2. Standard Value: The value from a reference standard or calibration standard.
  3. Tolerance: The acceptable range of error for the instrument.
  4. Full-Scale Range (FSR): The maximum range of the instrument.
  1. Error: The difference between the instrument reading and the standard value.
  2. Percentage Full-Scale Error: The error expressed as a percentage of the full-scale range.
  3. Pass/Fail: Whether the instrument is within the acceptable tolerance.
Calibration Verification Calculator 2
Calibration Verification Calculator 3

Check if the absolute value of the percentage full-scale error is within the specified tolerance.

Click here for Calibration Error Calculator for Instruments

Let’s perform a calibration verification example for a pressure transmitter that operates within a negative to positive range. We’ll assume the following details:

  • Full-Scale Range (FSR): -50 psi to 50 psi
  • Standard Values (Reference Pressure): -50, -25, 0, 25, 50 psi
  • Instrument Readings: -49, -24, 1, 24, 51 psi
  • Tolerance: ±1% of full scale (1 psi)
  • Full-Scale Range (FSR): 100 psi (since -50 to 50 psi gives a range of 100 psi)

Click here for Online Calibration Test Points Value Calculator

  • Standard Value: -50 psi
  • Instrument Reading: -49 psi
  • Error: −49−(−50)=1 psi
  • Percentage Full-Scale Error: (1/100​)×100=1%
  • Pass/Fail: Pass (within tolerance)
  • Standard Value: -25 psi
  • Instrument Reading: -24 psi
  • Error: −24−(−25)=1 psi
  • Percentage Full-Scale Error: (1/100)×100=1%
  • Pass/Fail: Pass (within tolerance)
  • Standard Value: 0 psi
  • Instrument Reading: 1 psi
  • Error: 1−0=1 psi
  • Percentage Full-Scale Error: (1/100)×100=1%
  • Pass/Fail: Pass (within tolerance)
  • Standard Value: 25 psi
  • Instrument Reading: 24 psi
  • Error: 24−25=−1
  • Percentage Full-Scale Error: (−1/100)×100=−1%
  • Pass/Fail: Pass (within tolerance)
  • Standard Value: 50 psi
  • Instrument Reading: 51 psi
  • Error: 51−50=1 psi
  • Percentage Full-Scale Error: (1/100)×100=1%
  • Pass/Fail: Pass (within tolerance)
Standard Value (psi)Instrument Reading (psi)Error (psi)Percentage Full-Scale Error (%)Pass/Fail (±1%)
-50-4911Pass
-25-2411Pass
0111Pass
2524-1-1Pass
505111Pass

The calibration verification of the pressure transmitter is successful in this case because every data point falls within the designated full-scale error tolerance of ±1%. 

By using this method the accuracy of the instrument is verified throughout its whole operating pressure range from negative to positive providing accurate readings for crucial applications.

Click here for More instrumentation calculators

Sundareswaran Iyalunaidu

With over 24 years of dedicated experience, I am a seasoned professional specializing in the commissioning, maintenance, and installation of Electrical, Instrumentation and Control systems. My expertise extends across a spectrum of industries, including Power stations, Oil and Gas, Aluminium, Utilities, Steel and Continuous process industries. Tweet me @sundareshinfohe

Related Articles

Back to top button

Adblock Detected

We Noticed You're Using an Ad Blocker Hi there! We understand that ads can be annoying, but they help support our website and allow us to continue providing you with high-quality content. Please consider whitelisting our site or disabling your ad blocker while you visit. Your support means a lot to us! Thank you for understanding!