Instrument Calibration Verification Calculator
An instrument calibration verification calculator confirms that instruments provide accurate and precise measurements by comparing their performance to a predefined standard.
Calibration Verification Calculator
The following is a thorough example of a pressure transmitter calibration with five points.
Calculator Inputs
- Instrument Reading: The value provided by the instrument being tested.
- Standard Value: The value from a reference standard or calibration standard.
- Tolerance: The acceptable range of error for the instrument.
- Full-Scale Range (FSR): The maximum range of the instrument.
Calculator Outputs
- Error: The difference between the instrument reading and the standard value.
- Percentage Full-Scale Error: The error expressed as a percentage of the full-scale range.
- Pass/Fail: Whether the instrument is within the acceptable tolerance.
Formula and Calculation Steps:
Step 1: Error Calculation
Step 2: Percentage Full-Scale Error Calculation
Step 3: Pass/Fail Determination
Check if the absolute value of the percentage full-scale error is within the specified tolerance.
Click here for Calibration Error Calculator for Instruments
Example Calculation for a 5-Point Calibration of a Pressure Transmitter:
Let’s perform a calibration verification example for a pressure transmitter that operates within a negative to positive range. We’ll assume the following details:
- Full-Scale Range (FSR): -50 psi to 50 psi
- Standard Values (Reference Pressure): -50, -25, 0, 25, 50 psi
- Instrument Readings: -49, -24, 1, 24, 51 psi
- Tolerance: ±1% of full scale (1 psi)
Step-by-Step Calculation:
Full-Scale Range:
- Full-Scale Range (FSR): 100 psi (since -50 to 50 psi gives a range of 100 psi)
Click here for Online Calibration Test Points Value Calculator
Calibration Point 1:
- Standard Value: -50 psi
- Instrument Reading: -49 psi
- Error: −49−(−50)=1 psi
- Percentage Full-Scale Error: (1/100)×100=1%
- Pass/Fail: Pass (within tolerance)
Calibration Point 2:
- Standard Value: -25 psi
- Instrument Reading: -24 psi
- Error: −24−(−25)=1 psi
- Percentage Full-Scale Error: (1/100)×100=1%
- Pass/Fail: Pass (within tolerance)
Calibration Point 3:
- Standard Value: 0 psi
- Instrument Reading: 1 psi
- Error: 1−0=1 psi
- Percentage Full-Scale Error: (1/100)×100=1%
- Pass/Fail: Pass (within tolerance)
Calibration Point 4:
- Standard Value: 25 psi
- Instrument Reading: 24 psi
- Error: 24−25=−1
- Percentage Full-Scale Error: (−1/100)×100=−1%
- Pass/Fail: Pass (within tolerance)
Calibration Point 5:
- Standard Value: 50 psi
- Instrument Reading: 51 psi
- Error: 51−50=1 psi
- Percentage Full-Scale Error: (1/100)×100=1%
- Pass/Fail: Pass (within tolerance)
Table with Full-Scale Error Data
Standard Value (psi) | Instrument Reading (psi) | Error (psi) | Percentage Full-Scale Error (%) | Pass/Fail (±1%) |
-50 | -49 | 1 | 1 | Pass |
-25 | -24 | 1 | 1 | Pass |
0 | 1 | 1 | 1 | Pass |
25 | 24 | -1 | -1 | Pass |
50 | 51 | 1 | 1 | Pass |
The calibration verification of the pressure transmitter is successful in this case because every data point falls within the designated full-scale error tolerance of ±1%.
By using this method the accuracy of the instrument is verified throughout its whole operating pressure range from negative to positive providing accurate readings for crucial applications.
Click here for More instrumentation calculators