Instrument Calibration Verification Calculator
An instrument calibration verification calculator confirms that instruments provide accurate and precise measurements by comparing their performance to a predefined standard.
The following is a thorough example of a pressure transmitter calibration with five points.
Calibration Verification Calculator 1
  1. Instrument Reading: The value provided by the instrument being tested.
  2. Standard Value: The value from a reference standard or calibration standard.
  3. Tolerance: The acceptable range of error for the instrument.
  4. Full-Scale Range (FSR): The maximum range of the instrument.
  1. Error: The difference between the instrument reading and the standard value.
  2. Percentage Full-Scale Error: The error expressed as a percentage of the full-scale range.
  3. Pass/Fail: Whether the instrument is within the acceptable tolerance.
Calibration Verification Calculator 2
Calibration Verification Calculator 3
Check if the absolute value of the percentage full-scale error is within the specified tolerance.
Click here for Calibration Error Calculator for Instruments
Let’s perform a calibration verification example for a pressure transmitter that operates within a negative to positive range. We’ll assume the following details:
  • Full-Scale Range (FSR): -50 psi to 50 psi
  • Standard Values (Reference Pressure): -50, -25, 0, 25, 50 psi
  • Instrument Readings: -49, -24, 1, 24, 51 psi
  • Tolerance: ±1% of full scale (1 psi)
  • Full-Scale Range (FSR): 100 psi (since -50 to 50 psi gives a range of 100 psi)
Click here for Online Calibration Test Points Value Calculator
  • Standard Value: -50 psi
  • Instrument Reading: -49 psi
  • Error: ?49?(?50)=1 psi
  • Percentage Full-Scale Error: (1/100?)×100=1%
  • Pass/Fail: Pass (within tolerance)
  • Standard Value: -25 psi
  • Instrument Reading: -24 psi
  • Error: ?24?(?25)=1 psi
  • Percentage Full-Scale Error: (1/100)×100=1%
  • Standard Value: 0 psi
  • Instrument Reading: 1 psi
  • Error: 1?0=1 psi
  • Standard Value: 25 psi
  • Instrument Reading: 24 psi
  • Error: 24?25=?1
  • Percentage Full-Scale Error: (?1/100)×100=?1%
  • Standard Value: 50 psi
  • Instrument Reading: 51 psi
  • Error: 51?50=1 psi
Standard Value (psi)Instrument Reading (psi)Error (psi)Percentage Full-Scale Error (%)Pass/Fail (±1%)
-50-4911Pass
-25-2411Pass
0111Pass
2524-1-1Pass
505111Pass
The calibration verification of the pressure transmitter is successful in this case because every data point falls within the designated full-scale error tolerance of ±1%.
By using this method the accuracy of the instrument is verified throughout its whole operating pressure range from negative to positive providing accurate readings for crucial applications.
Click here for More instrumentation calculators