Calibration

Step-by-Step Procedures for Pressure Gauge Calibration

What is a Pressure Gauge?

A self-contained indication that converts the measured process pressure into the mechanical motion of a pointer is what is frequently represented by the word “pressure gauge.”

pressure gauge calibrated 1

An instrument for measuring the pressure of a fluid or gas in a system is a pressure gauge. Typically, it comprises of a circular dial with a moving needle or pointer that responds to changes in pressure and a device that converts those changes in pressure into movement of the needle.

Monitoring the pressure in pipes, boilers, tanks, and other pieces of equipment are just a few of the many industrial, commercial, and domestic applications where pressure gauges are often utilised.

What are the Four 4 Types of Pressure Gauge?

  • A curved tube is used in a Bourdon tube gauge, which expands or contracts when pressure varies, causing the needle to move.
  • A flexible diaphragm is used in a diaphragm gauge, which moves in reaction to variations in pressure.
  • A flexible metal bellows is used in bellows gauges, which expand or contract in response to variations in pressure.
  • Digital gauge: This kind of gauge measures pressure using electronic sensors and shows the findings on a digital screen.

How do you Calibrate a Pressure Gauge?

7 Steps Calibration Procedure for Pressure Gauge

This procedure provides a thorough explanation of how to calibrate a pressure gauge in the workshop using comparator.

Step 1: Prepare the Tools Required for Pressure Gauge Calibration

The tools needed to calibrate a pressure gauge will depend on the setup and method used to calibrate it. But here are some general tools that are often used to calibrate pressure gauges:

pressure gauge calibrated2
 pressure gauge calibrated 3
  • Reference standard digital pressure gauge.
  • Pressure comparator.
  • Mineral oil for comparator.
  • Sprit level.
  • Pointer puller or pointer remover.
  • Necessary hand tools.
  • Standard adaptor fittings with seals.
  • Soft Cloth for cleaning.

Step 2: Safety Precautions

  • Please click on the link below for details on process industry calibration procedures, general safety precautions, and basic safety information – Basic Safety and General Considerations for Process Industries Calibration Process
  • Locate the pressure gauge you want to remove. Check to make sure it’s the right converter, and record any pertinent information Tag number (e.g., the manufacturer, model number, pressure range, etc.).
  • Depending on the kind of system, you may need to depressurize it before detaching the pressure gauge. Use the right procedures to depressurize the system (e.g., vent the pressure or isolate the system).
  • Stop the process and release any pressure or fluid that may be present in the system before disconnecting the pressure gauge.
  • Use the built-in valve or bleed port on the pressure gauge to gradually release pressure. This will help in avoiding sudden pressure releases that can result in harm.
  • Carefully remove the tubing connections from the pressure gauge. Observe all necessary safety procedures, such as wearing insulated gloves or other protective gear, and take care not to damage the tubing or fittings.
  • The process tapping has to be quickly caped or sealed once the gauge is disconnected to stop any fluid or gas from escaping.
  • To stop the pressure gauge from dropping or causing any risks, secure it. Any protective coverings, brackets, or hardware that were taken off during installation may need to be placed back on.
  • Mark or tag the disconnected gauge with the necessary details, including the time, the cause of removal, and any other relevant information. This will help keep the gauge from being confused or misidentified in the future.
  • Follow all relevant lockout/tagout procedures to prevent an unintended start. Make sure the pressure gauge is completely separated from the process.
pressure gauge calibrated 4
  • To calibrate a pressure gauge, you need a setup that makes sure the process is accurate and consistent.
  • Choose a reference standard that will be suitable for the calibration. This could be a pressure comparator, calibrator and another gauge whose accuracy is known.
  • The reference standard pressure gauge should be correct for the range of the gauge being calibrated and also it should be 10 times more accurate than the pressure gauge under calibration.
pressure gauge calibrated 5
  • The comparator base should be placed on a flat surface.
  • Check the four hand pump wheels’ spokes in the comparator to see whether they are securely fastened.
pressure gauge calibrated 12
  • Use the provided spirit level to ensure the piston-cylinder system’s comparator system base is level. 
pressure gauge calibrated 6
  • Use the leveling screws to modify the slope as required.
  • Use a bonded seal to secure the junction where the suitable connection is attached to the gauge stand, and then screw a test gauge (for installation, use a known gauge) into place.
  • Try to pull the plug out of the reservoir to remove the filler plug. (This plug should be left out while the device is being used.)
pressure gauge calibrated 7
  • Open both the A and B valves.
  • The valve in the rear is called “Valve A,” and it is used to control the flow from the spindle pump’s ram with the bigger diameter. The front valve is called “Valve B,” and it controls how much liquid goes into and out of the reservoir.
  • Completely clockwise turn the spindle pump handle.
pressure gauge calibrated 8
  • Fill the tank with the correct liquid. Use the oil that came with the system or a substitute that has been approved. Do not use other liquids.
  • Turn the pump’s handle all the way anticlockwise.
  • If required, top out the reservoir.
  • The calibration setup for the pressure gauge is now ready to be used.

Step 4 : Calibration of Pressure Gauge

  • Use a clean, dry cloth to wipe the gauge to get rid of any loose dirt or debris. If required, use a cleaning solution that the manufacturer suggests. Do not use brushes or cleaning agents that might scratch the gauge face.
  • Moreover, clean the pressure gauge’s Bourdon tube in accordance with the manufacturer’s instructions to prevent the mixing of process fluid and oil in the comparator.
  • Look closely for any evidence of damage on the gauge, such as cracks or dents. Before calibrating, it may need to be fixed or replaced if you find any damage.
  • To dry the gauge if it has any moisture, use a fresh, dry towel. Prior to calibration, make sure the gauge is totally dry.
  • By utilizing the proper adapters and fittings, you may connect the pressure gauge to be calibrated and reference standard pressure gauge with oriented for viewing. Make sure the connections are secure and free of leaks.
  • Close the Valve B and keep Valve A open.
  • The comparator pump handle should be turned clockwise. By looking at both gauges, you can now see how the pressure will build within the comparator.
  • Each calibration point is controlled by rotating the pump spindle.
  • Use the reference standard to regulate the pressure when you apply a test pressure to the gauge. Record the measurements after allowing the gauge and reference standard to settle. 
  • Continue this procedure at several test pressure points on the calibrating gauge. Keep track of each pressure’s measurements.
  • Determine the difference between the reading from the gauge and the reading from the reference standard for each pressure. Using any calibration adjustments on the gauge, adjust it as needed to reduce the inaccuracy.
  • Remove the ring, window, gasket, and pointer to do adjustments.
pressure gauge calibrated 10
  • According to the dial, mark the case flange at 0, 25, 50, 75%, and 100% of the pressure gauge’s range.
  • Followed by that remove dial from the gauge 
  • Pointer should be gently pressed against the pinion and precisely aligned with the 0% mark on the case flange.
  • Start by applying pressure, then do the calibration.
pressure gauge calibrated 9
  • Adjustments for zero, span, and linearity error varied depending on the manufacturer and type of the pressure gauge. To learn about the calibration process and adjustments, always read the pressure gauge manual.
  • After making the necessary adjustments, remove the pointer away once again. Put in place the back dial, pointer and close the pressure gauge with case cover.
  • Apply the test pressure once again and compare the results after making any necessary adjustments. Continue doing this until the gauge’s readings fall within a reasonable tolerance of the reference standard.

Step 5: Recording Calibration

  • Do linearity tests in the upscale and downscale directions at 0%, 25%, 50%, 75% and 100% to make sure the pressure gauge is producing the correct numbers.
  • If the output value does not fall within an acceptable range, calibration is required. Once again, if the displayed values have departed from the permitted range, a pressure gauge must be repaired or replaced.
  • No extra calibration is required if every output value (+/- %) falls within acceptable bounds.
  • The output data should be entered into the as found/as left column of the blank calibration report.

Step 6: Completion of Calibration

  • When the calibration has been successfully conducted, attach the calibration label to the pressure gauge.
  • When the calibration is complete, clean the instrument, store it somewhere safe, and record the calibration data for later use.
  • Disconnect the pressure gauge, comparator, and other setup.
  • Reinstall the pressure gauge in the processing area.
  • Make sure the workspace is tidy.
  • De-isolate the equipment.
  • Resume the pressure gauge and check that it is functioning correctly.

Step 7: Calibration Report Preparation

The next picture shows that the pressure gauge sample report of calibration was done in workshop with pressure comparator and standard pressure gauge as the reference.

pressure gauge calibrated 11

The calibration process for pressure gauges complies to ISO standards, ensuring accurate measurements. Our method incorporates NIST principles for increased reliability, and compliance with ASME standards ensures performance integrity in a wide range of industrial applications. Follow our calibrated procedure for accurate readings, optimizing performance in manufacturing processes.

Sundareswaran Iyalunaidu

With over 24 years of dedicated experience, I am a seasoned professional specializing in the commissioning, maintenance, and installation of Electrical, Instrumentation and Control systems. My expertise extends across a spectrum of industries, including Power stations, Oil and Gas, Aluminium, Utilities, Steel and Continuous process industries. Tweet me @sundareshinfohe